Ana82
02/03/2012 6:15 a.m.
Yesterday I had a giant bubble in the catheter very stuck at the exit of the reservoir.
And what I did to remove it (this starting starts me doubts: Oops :) It was to disconnect, put a bolus and wait for it to come out.
I have done wrong?
Do you have to do something else?
Thank you.
No signature configured, add it on your user's profile.
Velia
02/03/2012 5:30 p.m.
Hello Ana, I see that you use the Medtronic bomb, when you have bubbles instead of a bolus it is more accurate to use the barley, or the reservoir option + equ.Infusion, depending on the model ... you rewind the pump and make a barley of the cable until the entire bubble disappears.
Greetings.
De los buenos tiempos, siempre quiero más...
Mamá de Ángela, ¡16 añitos, fiera!. Debut: octubre de 2003.
Bomba insulina Medtronic Paradigm Veo desde junio 2005
Última hemo 6.1
Ana82
02/06/2012 5:51 a.m.
Yes, Velia, I use Medtronic's little girl.
Thank you so much!!!!!
No signature configured, add it on your user's profile.
Hello Ana:
What you should do is select "Reservoir + Infusion team" to rewind and tell you why.
If you remove the air by putting a bolus, the bomb understands it as you have put it in your body and write it down as an active insulin.So, when you really want to put on a bolus because you go to eat or because you are high, she will understand that you have a lot of active insulin of that supposed bolus and will not correct you well.
I hope I have explained well.Chao!
No signature configured, add it on your user's profile.
Ana82
02/08/2012 5:56 a.m.
If Pepita you are right but of course !!!!
At the moment and without knowing it is the only thing that occurred to me.
Now I know you have to do everything as if it were a change but without changing the reservoir
Thank you so much!!!!!!!!!
No signature configured, add it on your user's profile.